OSEM reconstruction for SPECT using SIMIND

Hello, everyone.
I have a trouble, about OSEM reconstruction.
I was doing OSEM with pytomography, but the image after reconstruction dose not work. I was using a cylindrical homogeneous phantom with a diameter of 18cm, but the image is smaller when reconstructed.
I also do scatter correction and attenuation correction using CT images.

What are the possible causes of this and what is the solution?
Please help me.

SIMIND_image_bold
Projection image

Rec_image_bold
Reconstruction image

@Haruki which version of SIMIND are you using? There’s an issue in SIMIND v8.0 (where collimator dimensions are now expressed in mm instead of cm) that messes with PyTomography

Dear Luke.

Thank you for your reaction.
This was exactly the problem And I was able to resolve it.

One question: what do the units mean after reconstruction?
Is it [counts/MBq/sec] as in projection data?